Glenn wrote:
> Larry
> I'm not certain about the 856X series but I suspect it is the same as
> for the newer analyzers. If so, the way the detector works depends on a
> number of factors:
> o how fast one is sweeping.
> 0 detection mode (peak,sample, neg. peak)
> 0 resolution bandwidth
>
> As the analyzer's LO traverses the spectrum between measurement bins
> (each bin of width = span/#measurment points), depending on how the
> detector is selected, the A/D takes 1-to-many measurements. Maximum
> sweep rate is limited by the need to let the detected signal settle (the
> group delay of the RBW filter) and the ability of the LO itself to move
> quickly and accurately. For the ESA series, multiple measurements may
> be made within each bin. Peak Detect modes just save the max or min
> value for the bin. Sample Detect (probably) measures the value at the
> center(or end, I 'm not sure) of the bin and throws away any other
> measurment. For some analyzers there is a non-sweeping mode wherein the
> A/D runs as fast as it can. I think that is a 50 ns measurement time for
> ESA.
Thanks very much for the info Glenn, it is very hlepful.
> While there may be no dwell time selection, you may be able to influence
> the time spent by judicious selection of span, RBW and sweep time. But
> then, the question arises "why do you want to do that?".
I have to perform some EMI emissions tests that have specific dwell
time requirments
(>= 0.15 sec), and I will need to prove to my customer that I am
meeting these.
Thanks!
-larry
> The SA should
> already manage things to avoid over-sweeping, trying to take data faster
> than the if filters can handle, and the detection mode selection should
> allow you some fleximbility as to whether you're looking at a peak or
> sampled mode.
>
> Glenn
>
> > I am trying to determine what the dwell time is for these instruments.
> > By dwell time I mean the amount of time it samples the signal for each
> > data point. This does not appear to be a configurable value, nor does
> > there seem to be a way to query this value.
> >
> > For these instruments, is this parameter fixed or variable? If it is
> > fixed how
> > do I find out what it is? If it is variable, is it calcualted by simply
> > dividing the
> > sweep time by the number of data points?
> >
> > TIA,
> > -larry
> >
|